Problems During Observing .. And What to Do
1. Detector timeouts
Symptom: Count-down timer in NMSL GUI turns red keeps counting down negative numbers (to -105 or so), and no new images are read.
Frequency: Several times per night in the early Feb 2016 run.
Fix:
- Restart the controllers
ditscmd nohs nohs_endobs nocs nmsl reset
- Take a zero image, make sure it rads out, and check the image by running "mscstat"
- Create and start a new observing script
2. Images show bad amp
Symptom: One amp has extremely large noise, looks awful the display of the image with "mcsdisplay" with a noise > 1000 ADU/pix from the "mscstat" command.
Frequency: Several times in Dec 2015 run, never in early Feb 2016 run.
Fix: Reset the NOCS controller by typing the following in a NOCS xterm window:
nocs reset ccp nocs init ccp
Then take two zero images. The first image is always bad. The second should be OK. Check it using "mscstat <filename>" once it has read out. All rms values should be between 4-7 ADU except for amp 6 with will be 8-10 ADU.
3. 4MAPS primary mirror support goes “OFF AIR” during observing
Symptom: The OA should notice when this happens, but you can also see if the mirror support is listed as "OFF AIR" on the 4MAPS GUI. The images will also appear to have very large and strangely-shaped PSFs.
Frequency: This happened several times per night in the Dec 2015 run, but never in the early Feb 2016 run.
Fix:
- Type Control-C to exit out of the observing script
- Wait for it to finish and for “DONE” to appear on the NMSL GUI
- Type “ditscmd nohs nohs_endobs” in the xterm window in which the script was stopped. This should readout the last image and display it on the real-time display ximtool window
- Tell the OA they can move the telescope to zenith and reset 4MAPS
- Check the last image taken when 4MAPS failure occurred - the image quality is probably bad. Add this as a bad exposure in the file ~products/mosaic3/logs/bad_expid.txt
- Create a new observing script
- Once the OA has reset 4MAPS, start the new script
4. Shifted images
Symptom: Sub-blocks of an image have been shifted in the readout, and is obvious on the display as the overscan region being in a different place. The readout system detects when this happens, adds a set of PIXCNT* cards in the image header, and a red line shows up on the obsbot/copilot strip chart. Every now and then, an image appears which shows a shifted region in one or more CCDs. Please log this. You can identify images with shifts by running
Frequency: A couple of times per night in the early Feb 2016 run.
Fix:
- Add this as a bad exposure in the file ~products/mosaic3/logs/bad_expid.txt
5. Repeat images
This is very hard to diagnose, and requires you to pay attention to the obsbot/copilot output, the individual images, and occasionally running the command:
hselect mos3*fits[0] $I,READTIME yes
in the IRAF window.
If the READTIME values begin to repeat, this is a sign that new images are not being written out. Another clue is that the read out counter on the NMSL gui goes straight from 44 to 0. A third clue is that the MOSSTAT command is failing to find any stars in the image, even though the offsets were small when the sequence was started.
If any of these are true, then it means that the images are just duplicates of previous images. You need to stop the sequence using CTRL-C in the window where the script is being exectuted, and restart everything:
- Stop MOSAIC (from the startup gui)
- In a NOCS xterm window, type:
nocs nuke pana nocs nuke dhs
- Start MOSAIC (from the startup gui)
Once everything comes up and you have rearranged windows how you like, take a ZERO frame to make sure that the amps are all OK. Check the noise on the zero image using mscstat. If this is not OK, then see instructions for Problem 1.
From here, it is best to make a new plan (strategy file) and start it up.
Log all the images that had identical READTIME values.
6. Focus drifting
If you lose track of the focus and don’t know whether you are below or above the focus value, wait for the readout countdown to start and then type CTRL-c in the window executing the script. Wait for the readout to complete and the “DONE” to appear on the NMSL window. Then execute
ditscmd nohs nohs_endobs
Then do a focus sequence.
- Create a focus script using the “Focus” button on the NGUI This pops up the following window:
- use a good guess for the mid-point:
focus in zd filter = -8400 + (1.4C - Ttruss)*110
- an instrument focus step of -100 (or -70, if the seeing is good)
- at least 7 focus iterations and
- a detector pixel shift of 30;
- Make sure you have chosen the zd filter
- execute the focus sequence
- Edit the /data2/observer/mscfoc.cl script to change the file name of the focus image
- In the IRAF window, type "mscfoc.cl" and follow the instructions
- Once you have determined the best focus:
- set the focus value in the MCCD window and hit “Apply”
- log the Truss temperature and focus values.
- Create a new observing script
- Tell the OA that you have determined a new focus and changed the focus.
- Restart observing
7. Images of stars on the guider are bouncing around
- If the wind is high, ask the OA to close the louvers
Attachments (7)
-
4MAPS_gui.tiff (59.0 KB) - added by 9 years ago.
Screen shot of 4MAPS display window
-
NOCS_FOCUS.tiff (37.5 KB) - added by 9 years ago.
Image of NOCS FOCUS gui
-
4MAPS_OFF_AIR.tiff (61.7 KB) - added by 9 years ago.
4MAPS gui showing OFF AIR
-
windshake.tiff (11.8 KB) - added by 9 years ago.
Image of star with windshake
-
nocs_timeout.png (22.2 KB) - added by 9 years ago.
NOCS window timeout message
-
imageshift.png (306.4 KB) - added by 9 years ago.
example of a shifted image
- pupilghost_obscura_137456.jpg (56.1 KB) - added by 7 years ago.
Download all attachments as: .zip